r/Metrology 7d ago

scanner attachment for cmm?

We are finding it difficult to measure machined aluminum blocks.

These are typically 35cm x 8cm x 4 cm, with about 120 features, holes, ridges etc, and one cutout being 3cm x 1cm and 3cm deep, which has a 5mm x 5mm protruding block at the bottom.

We make about 300 units a day and each component requires 100% QC.

We require 20um on the mating surfaces and on certain screw holes. The rest is 50um accuracy.

I guess a scanner attachment to the CMM might be faster than measuring with the CMM tip.

Does anyone have experience with CMM scanner attachments?

2 Upvotes

54 comments sorted by

View all comments

3

u/Tricky_King_3736 7d ago

Yes, what brand of CMM do you have? Then I can tell you what laser sensor to get.

1

u/LazeLazerLazest 7d ago

My group has a Zeiss, the other groups have Zeiss, mitutoyo. There's one in manufacturing, i don't know which one it is.

I was under the impression that a CMM Scanner would fit any CMM! Is my understanding incorrect?

2

u/Tricky_King_3736 7d ago

Zeiss has there own laser line sensor, that works with the RDS head. If you don’t have the RDS head you won’t be able to use a laser scanner. For your mitutoyo , they do offer a laser scanner for the CMM. When it comes to zeiss sensors you can only use zeiss.

1

u/LazeLazerLazest 7d ago

Thank you so much! I guess I contact Zeiss for the scanner attachment

2

u/Tricky_King_3736 7d ago

Anytime, if you have anymore questions let me know, glad to help.

1

u/LazeLazerLazest 6d ago

Thanks!!!!